I have a Verus Pro and an eepv 302 AL Snap on transducer. My question is what does everyone else get for a maximum reading for vacuum? It says it should read to 20 inHg. I thought the one I had was bad so I had it warranteed. The new one is worse. My old one would read 14.4 max and the new one reads 13.9 max. In both cases I used a vacuum pump with a guage plus a separate vacuum guage to double check. I can max the pump out to 22inHg and the max reading is 13.9inHg. Yes I calibrate the transducer before every test. Shouldn't it read at least 20inHg if that is supposed to be the range?
Thanks and Merry Christmas to all
Thanks and Merry Christmas to all
they are sold as pressure transducers, and seem to display pressure referenced to atmospheric, they dont to be displaying absolute pressures, and they dont list an actual definitive vacuum spec that I see. Its sold as a 0-100psig sensor. Id consider the vacuum a nice bit if overrun but wouldnt rely on it for diagnosis, and instead pickup either a Fluke one that actually lists precise specs for pressures below 0psig or assemble my(your) own from commercially available transducers from Omega, Johnson Controls and the like.